Cross-check @ 40 bar recommended. TF asks for an additional PQR test @ 10 MPa/200 °C.
In the cross-check the leak rates at low stresses are a little bit higher, at higher stresses they are lower than in the double tst of IDT. But all leak rates are in the acceptable range.
The PQR value at 10 MPa/200 °C in the cross-check confirms the value of the double test (0,55 vs. 0,50).
After performing the cross-checks with positive results the datasheet of IDT is accepted.
In the cross-check the leak rates at low stresses are a little bit higher, at higher stresses they are lower than in the double tst of IDT. But all leak rates are in the acceptable range.
The PQR value at 10 MPa/200 °C in the cross-check confirms the value of the double test (0,55 vs. 0,50).
After performing the cross-checks with positive results the datasheet of IDT is accepted.